Originally Posted by JasonJayhawk This would be an interesting question to call and ask one of their reps. Hmmmm, a recorded phone call...to be posted on the forum!
Jedi...did you ever call your local rep? |
No, because once I found the manual at.... Product User Guides and then clicking on... http://www.disetronic-usa.com/files/...2%20121605.pdf and then going to pages....151....
Alert A5: PUMP TIMER
Your pump’s operating time will expire soon. It is recommended that you replace your pump after a period of four years of running time. After the alert A5: PUMP TIMER occurs, the ACCU-CHEK Spirit insulin pump will display the symbol to
remind you to check the timer.
1. Press ftwice to turn off and to confirm the alert.
2. Put your pump in RUN mode.
3. Go to the INFORMATION screen on your pump.
4. Move to the PUMP TIMER screen.
The days until the pump timer expires are shown.
When the PUMP TIMER reaches zero, the pump will change to STOP mode and no longer operate. An error, E5: END OF OPERATION, will appear on the display. Contact Pump Support at 1-800-688-4578 to discuss your options for further
use of your ACCU-CHEK Spirit insulin pump. See the section “Review the Time Remaining” (page 112) for more information.
Note:
The alert A5: PUMP TIMER ALERT and the error E5: END OF OPERATION are based on running time (the total time the pump is in RUN mode) and not calendar years. However, the warranty is based on the date the pump was purchased.
From page 158.....
Error E5: END OF OPERATION
Your pump timer has reached zero. The pump will change to STOP mode and will no longer operate. An error E5: END OF OPERATION will appear on the display. It is recommended that you replace your pump after four years. Before this error
occurs, an alert A5: PUMP TIMER will remind you to replace your pump soon.
1. Press ftwice to turn off and to confirm the error.
2. Contact Pump Support at 1-800-688-4578 to discuss your options for further use of the ACCU-CHEK Spirit insulin pump.
And above said to go to page 112...so...
Review the Time Remaining
Your pump is designed to be a highly reliable system that will give you worry-free performance for a long time. For best performance, a pump timer limits the operating time of the pump. This timer counts the number of days of operation
remaining for the ACCU-CHEK Spirit insulin pump.
Before the pump timer expires, an alert will remind you of the upcoming end of operating time so that you can arrange to get a new one. The pump timer symbol ( ) appears in the RUN screen or STOP screen as a reminder. When the timer expires, an error occurs and your pump goes into STOP mode. It can no longer be put into RUN mode. See the sections “Alert A5: PUMP TIMER” (page 151) and “Error E5: END OF OPERATION” (page 158) for more information. Review the number of days left until the pump timer expires in the INFORMATION menu.
Follow the steps below:
1. Press dto move to the INFORMATION screen. Press fto select.
2. Press dto move to the PUMP TIMER screen. The days until the pump timer expires is shown.
3. Press fto exit.
3 times in their manual it mentions shutting off. And this is from the Disetronic USA site not the Disetronic DE site. I think that explains enough that it shuts off, so why even put it in their manual if it doesn't?