You are currently viewing our boards as a guest which gives you limited access to view most discussions and access our other features. By joining our free community you will have access to post topics, communicate privately with other members (PM), respond to polls, upload content and access many other special features.

Registration is fast, simple and absolutely free so please, join our community today!

If you have any problems with the registration process or your account login, please contact contact us.


Reply
This 15% rule is just plain ridiculous LinkBack Thread Tools Display Modes
  #1 (permalink)  
Old 03-26-2008, 07:38 AM
Member
I am a: Type 1
 
Join Date: Mar 2008
Location: South Carolina
Posts: 115
Thumbs down This 15% rule is just plain ridiculous

I know the guidelines for BG Meters is that are required to by accurate +/- 15%. OK, I just took 3 readings in row, using different fingers. My readings were in this order:
READING A: 207
READING B: 179
READING C: 200

- OK, -15% Of the Reading A gives me 175.95. The other two readings fall within the accuracy guidelines.

- BUT, +15% of Reading B gives me 205.85 not within the accuracy guidelines.

- All three are within 15% of Reading C (-15% = 170 and +15%= 230)

I really wish the FDA would titghten down this 15% to at least 7.5% accuracy guideline.
__________________
Glucoweb
diagnosed 1995 pumping since 2000
Pump: Deltec Cozmo
Reply With Quote
  #2 (permalink)  
Old 03-26-2008, 07:41 AM
Scratch's Avatar
Senior Member
I am a: Type 1
 
Join Date: Oct 2006
Posts: 1,085
I thought it was +/- 20% which probably won't make you any happier regarding the somewhat diverse readings you had above, although the 207 and 200 would likely indicate your true glucose level is closer to that than 179.

I don't know. I think it's something people tend to fret too much over and in some ways I don't want to give the manufacturers any more potential reasons to jack up the costs on the testing strips.
__________________
MDI, Lantus and Novolog
A1c 10/08 -- 5.4%
A1c 4/08 -- 5.7%
A1c 8/07 -- 5.6%
A crazy rambling log/thread about getting fit
Scratch's running log
Reply With Quote
  #3 (permalink)  
Old 03-26-2008, 07:48 AM
morrisma's Avatar
Senior Member
I am a: Type 1
 
Join Date: Dec 2007
Location: Landenberg, PA
Posts: 1,337
Although I seriously doubt one meter is significantly better than any others, there are ample opportunities to try them out for free. I have seen at least one free meter offer a month and they always come with enough strips to do an accuracy test such as the one you describe. The downside of course is that you need to keep doing the test on every lot until you are sure they can (or cannot) meet some higher standard.

There may also be online results of more scientific testing on meter's websites and maybe independent sites as well.
Mike
__________________

Type 1 since '88
Pumping since 2002
Reply With Quote
  #4 (permalink)  
Old 03-26-2008, 08:35 AM
johgn's Avatar
Member
I am a: Type 1
 
Join Date: Dec 2006
Location: Baltimore, USA
Posts: 191
I was under the impression you would need to use the same finger each time for this test.
__________________
" ..That the human condition can be summed up in just one word, and this is the word: Embarrassment."
Kurt Vonnegut
Reply With Quote
  #5 (permalink)  
Old 03-26-2008, 08:51 AM
princesslinda's Avatar
Super Moderator
I am a: Type 2
 
Join Date: Dec 2006
Location: Knoxville, TN
Posts: 7,261
Quote:
Originally Posted by johgn View Post
I was under the impression you would need to use the same finger each time for this test.
Even using the same finger, same stick, I have yet to get 2 readings in a row that were exactly the same.
__________________
T2, diagnosed 8/31/06.
Byetta 5 mcg
HCTZ 12.5 mg every other day for BP
Enalapril 20 mg 1 daily (ace-inhibitor)
Lower carb dieter (approx. 75 total carbs/day, more on weekends), taking chromium, multivitamin and fish oil tablets


Initial A1C 8/06: 9.6
11/06: 6.2.
03/07: 5.3
06/07: 5.4
10/07: 5.3
05/08: 6.2 (right after dealing with shingles and bronchitis)


Reply With Quote
  #6 (permalink)  
Old 03-26-2008, 09:23 AM
xMenace's Avatar
Senior Member
I am a: Type 1
 
Join Date: Jun 2006
Location: Rothesay, New Brunswick Canada, eh
Posts: 7,140
Quote:
Originally Posted by princesslinda View Post
Even using the same finger, same stick, I have yet to get 2 readings in a row that were exactly the same.

I have! But only once, ever.

I am planning on running a test sometime of various meters using calibration solutions. My daughter is studying chemistry at university, and I'm hoping she'll prepare some glucose solutions at various levels for me to use. I'd like to know how close the very high and very low readings are. I may just talk to her prof or another one directly. It should be easy to do. I also notice that sanvita produces test solutions in three strengths. I may try to get some of those.
__________________
Michael Pollan on CBC

In Defense of Food with Michael Pollan


T1 1975, MM 722 pump

10/08
A1C 7/08 6.1%
HDL - 1.74 (67)
LDL - 1.89 (73)
Triglicerides - 0.52 (47.0)


7/08
A1C 7/08 5.9%
HDL - 1.55 (59.9)
LDL - 1.76 (68.1)
Triglicerides - 0.44 (40.0)

John
Reply With Quote
  #7 (permalink)  
Old 03-26-2008, 09:51 AM
Scrabblechick's Avatar
Senior Member
I am a: Type 2
 
Join Date: Mar 2008
Location: Alabama
Posts: 925
Good idea, X! Let us know what you come up with.

I've said before that I have two meters, a True Track and a One Touch Ultra2. The One Touch consistently runs about 5-10 points higher than the TT. When it's not 5 points lower. LOL. Yesterday, I did three consecutive tests with my TT and the results were within 5 points of each other. So, I'm pretty comfortable with its accuracy and consistency.

I just have to use one meter and stick with it, or my fingers would leak blood! LOL. My One Touch is my backup meter.

As long as my next bloodwork agrees with the results I've been getting on my TT, I'll be a happy camper!
__________________
Glycemic impact diet
exercise
Metformin 2000 mg
Byetta 10 mcg/2x daily
Enalapril 40 mg
A1C, 11-14-08: 5.2!!
A1C, 8-7-08: 6.3
A1C, 5-1-08: 5.6!!
A1C, 2-5-08: 7.4
Reply With Quote
  #8 (permalink)  
Old 03-26-2008, 09:56 AM
xMenace's Avatar
Senior Member
I am a: Type 1
 
Join Date: Jun 2006
Location: Rothesay, New Brunswick Canada, eh
Posts: 7,140
I always get my math wrong, but here's something to consider

(179 + 207) / 2 = 193 (estimated median)

193 * 1.15 = 222 (+15%)
193 * .85 = 164 (-15%)

179 / 193 = 93%
207 / 193 = 107%

Your numbers seem ok to me.
__________________
Michael Pollan on CBC

In Defense of Food with Michael Pollan


T1 1975, MM 722 pump

10/08
A1C 7/08 6.1%
HDL - 1.74 (67)
LDL - 1.89 (73)
Triglicerides - 0.52 (47.0)


7/08
A1C 7/08 5.9%
HDL - 1.55 (59.9)
LDL - 1.76 (68.1)
Triglicerides - 0.44 (40.0)

John
Reply With Quote
  #9 (permalink)  
Old 03-26-2008, 10:15 AM
Member
I am a: Type 1
 
Join Date: Mar 2008
Location: South Carolina
Posts: 115
Quote:
Originally Posted by xMenace View Post
I always get my math wrong, but here's something to consider
(179 + 207) / 2 = 193 (estimated median)
193 * 1.15 = 222 (+15%)
193 * .85 = 164 (-15%)
179 / 193 = 93%
207 / 193 = 107%
Your numbers seem ok to me.
I do not believe the quidelines discuss taking the median of multiple readings then make sure all the readings fall within 15% of the median. I did find some interesting information here:
Diabetes Information - Glucose Meters and Diabetes Management
I like the 1993 recommendation, too bad it was never followed and that was 15 years ago.
__________________
Glucoweb
diagnosed 1995 pumping since 2000
Pump: Deltec Cozmo
Reply With Quote
  #10 (permalink)  
Old 03-26-2008, 10:25 AM
Member
I am a: Type 1
 
Join Date: Jan 2008
Location: London UK
Posts: 275
one time i got a reading of 2.7 (around 47 in U.S units). i thought WEIRD, lowest ever reading and i dont feel low at all (normally i can tell around 4 that im starting to slide). so i re-tested and was 7.0. iv not had any problems with my meter since then and have tested it against other peoples meters that are the same and it has been about right. just seems a little odd though!
__________________
Lizzie
Reply With Quote
  #11 (permalink)  
Old 03-26-2008, 10:47 AM
Eddy's Avatar
Senior Member
I am a: Type 1
 
Join Date: Jan 2008
Location: Kansas, US
Posts: 1,097
Quote:
Originally Posted by Lizzie G View Post
one time i got a reading of 2.7 (around 47 in U.S units). i thought WEIRD, lowest ever reading and i dont feel low at all (normally i can tell around 4 that im starting to slide). so i re-tested and was 7.0. iv not had any problems with my meter since then and have tested it against other peoples meters that are the same and it has been about right. just seems a little odd though!
Heh. I had the exact opposite happen a couple days ago. I *knew* I was hypo, and guessed in the low 50s to high 40s. Checked... 147. Nope, not buying it. Checked again... 46.

If in doubt, consider trusting yourself. If you don't want to that... have a little sugar to be safe.

I once had a batch of 100 bad strips. I tested... off-the-meter low. Retested... same. I _thought_ the meter was wrong, but could not swear to it.[1] I chugged a good amount of juice and soda, then proceeded to determine what reality was.

[1] I'd been on the phone for an hour with a friend. I was very relaxed, which reduces my hypo sensitivity.
__________________
Eddy


DXed 2007/04 = advanced-stage DKA, A1c of 12.9%, and BMI of 21.3
post-DX A1c = 5.4% @ 2008/07; 5.2% @ 2008/04; 5.3% @ 2007/12; 5.3% @ 2007/08
c-peptide = 0.0% @ 2008/07
current BMI = 26.0 (86kg on 182cm); want to get back to 23-24
basal = NPH and Levemir, ~35U daily (I really should start a thread)
bolus = 1:15 I:C ratio; varying mix of aspart, human R, human N

not a low-CHO eater... not even close!
last updated 2008/11/03

Reply With Quote
  #12 (permalink)  
Old 03-26-2008, 01:37 PM
Member
I am a: Type 1
 
Join Date: Mar 2008
Location: South Carolina
Posts: 115
Does anyone know of a campaign or lobby for the companies to improve their product accuracy? I usually include a paragraph about it in my annual letters to my representatives seeking diabetes funding. But do not know of a specific push for it.
__________________
Glucoweb
diagnosed 1995 pumping since 2000
Pump: Deltec Cozmo
Reply With Quote
  #13 (permalink)  
Old 03-26-2008, 01:46 PM
Kim_in_TN's Avatar
Senior Member
I am a: Type 2
 
Join Date: Feb 2008
Location: Tennessee
Posts: 678
I only test once in approximately the same location each time. I can't afford to be frustrated with my readings. I take my first reading and go with that!
__________________



Type 2 - Dx 9/11/03 - A1c 8.3
2/01/08 - A1c 6.3
4/01/08 - A1c 5.9
7/25/08 - A1c 5.9

Current Meds - Janumet 50/1000 2x daily
Reply With Quote
  #14 (permalink)  
Old 03-26-2008, 04:39 PM
JediSkipdogg's Avatar
Senior Member
I am a: Type 1
 
Join Date: Sep 2005
Location: Cincinnati, OH
Posts: 8,369
Quote:
Originally Posted by Glucoweb View Post
I do not believe the quidelines discuss taking the median of multiple readings then make sure all the readings fall within 15% of the median. I did find some interesting information here:
Diabetes Information - Glucose Meters and Diabetes Management
I like the 1993 recommendation, too bad it was never followed and that was 15 years ago.
The 15% is from lab results. So if your lab blood reading was 200, then your meter is perfectly within error from 170-230. Big gap huh? Yeah, it sucks.

Quote:
Originally Posted by Glucoweb View Post
Does anyone know of a campaign or lobby for the companies to improve their product accuracy? I usually include a paragraph about it in my annual letters to my representatives seeking diabetes funding. But do not know of a specific push for it.
There was a while ago but meter companies claim that it's impossible to keep the meters affordable then and a larger sample of blood would be required. You want a very accurate meter, check out the Hemocue Glucose 201. The meter runs $400, uses 5 uL of blood (that's 17 times the Flash and 5 times the Ultra line) and strips are $135 for 100 not covered by any insurance that I'm aware of.

That's why they don't push for more accurate meters, it's just unreasonable in terms of blood sample required and price.
__________________
●Blue Ash, Ohio Police Dispatcher
●Type 1 diabetic for 25 years (11 months old)
●Animas pumper since December of 2002
~IR 1000 (Dec. 2002-Jan. 2005)
~IR 1200 (Jan. 2005 - ?)
●LifeScan OneTouch UltraSmart

Diabetes is an Art, NOT a Science. You must master the control by skills and not by knowledge alone.
Reply With Quote

Reply


Thread Tools
Display Modes
Posting Rules
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts

vB code is On
Smilies are On
[IMG] code is On
HTML code is Off
Trackbacks are On
Pingbacks are On
Refbacks are On


» Log in
User Name:

Password:

Not a member yet?
Register Now!

All times are GMT -7. The time now is 09:16 PM.

For Advertising:

1 2 3 4 5 6 7 8 9 10 11 12 13 14 15 16 17 18 19 20 21 22 23 24 25 26 27 28 29 30 31 32